 Can you slow the scan rate of VI Logger scans per minute

Can you set up a logging task with VI Logger to log 1 scan per minute?
I'm using a DaqCard AI-16E-4 and SCXI 1120 Module.


Tue, 12 Oct 2004 20:37:31 GMT  
 Can you slow the scan rate of VI Logger scans per minute
Hello,

The only parameter that can be changed while setting up a task in VI
logger is the "scans per second", which  corresponds to your DAQ board
(1 to 100KS/s).

Since VI logger acquires continuously, the minimum number of scans
that you can do per second is one; this is why you will have to at
least log 60 values per minute into the file.  Unfortunately, there is
no way to change this parameter since it really depends on the
hardware of the board, not the software.

If your application requires to log every minute, I would recommend to
use labview (AI OnePoint.vi).  With  this software you will have more
flexibility over the "writing times" and the results will be the same.

Hope this helps.  Keep up with the great work!
Have a nice day!



Fri, 15 Oct 2004 21:13:01 GMT  
 
 [ 2 post ] 

 Relevant Pages 

1. RELATION BETWEEN SCAN RATE AND SAMPLING RATE

2. Anti-Port scan.... (Port Monitor / Logger)

3. Scan (+Binary Scan) to Array feature addition?

4. Scan (+Binary Scan) to Array feature addition?

5. PCI6024E minimum scan rate

6. Simultaneously gathering multiple channels at different scan rates

7. Simultaneously gathering multiple channels at different scan rates

8. error 10401....but only for high scan rate

9. scan rate using hardware timing.

10. how to get a preview of actual AI scan rates of MIO cards

11. scan rate exceed the limit?

12. Scan codes and repeat rates

 

 
Powered by phpBB® Forum Software